Our full technical support staff does not monitor this forum. If you need assistance from a member of our staff, please submit your question from the Ask a Question page.


Log in or register to post/reply in the forum.

Email send and cell200 logger


PPeters Aug 21, 2018 04:59 AM

Hi,

I am having an issue with setting up a cr300 cell200 logger to do an exception report Email. I am using proven code that I can reliable test on a CR1000x platform using ethernet connection to a cybertec SMM-400 modem. I am using the same sim in both cases and we run a dedicated APN. 

using this line of code for both email sends

EmailSuccess = EmailSend (ServerAddr,ToAddr,FromAddr,Subject,Message,Attach,UserName,Password,EmailResult)

On both systems I "Ping" our mail sever prior to attempting the email send and in both cases get a successful response

however the CR300 is failing to deliver the message

using the W mode on terminal and looking at the cell modem traffic I get this response

16:45:55.26 T 7E ~
16:46:05.26 T 7E FF 7D 23 C0 21 7D 25 7D 22 7D 20 7D 37 50 65 ~.}#.!}%}"} }7Pe
16:46:05.26 T 65 72 20 6E 6F 74 20 72 65 73 70 6F 6E 64 69 6E er not respondin
16:46:05.26 T 67 2E 50 7E g.P~
16:46:08.26 T 7E FF 7D 23 C0 21 7D 25 7D 23 7D 20 7D 37 50 65 ~.}#.!}%}#} }7Pe
16:46:08.26 T 65 72 20 6E 6F 74 20 72 65 73 70 6F 6E 64 69 6E er not respondin
16:46:08.26 T 67 73 F9 7E gs.~
16:46:29.26 T 2B 2B 2B +++
16:46:33.77 T 41 54 48 0D 0A ATH..
16:46:44.88 T 41 54 44 2A 39 39 2A 2A 2A ATD*99***
16:46:44.88 T 31 23 0D 0A 1#..
16:47:05.49 T 41 54 44 2A 39 39 2A 2A 2A 31 23 0D 0A ATD*99***1#..
16:47:26.10 T 41 54 44 2A 39 39 2A 2A 2A 31 23 0D 0A ATD*99***1#..

with the modem rebooting??

not sure what is going on and what else I need to do to test further

I can forward full code and log file if this assists

was running 7.02 of firmware initially then 7.04 with no change

consistant issue over two loggers

regards

Paul


PPeters Aug 22, 2018 03:50 AM

https://www.dropbox.com/sh/fy6rv9ju7mh65mi/AAC7Iqi3flARoaOuGtfSe22Xa?dl=0

Ran some further tests and produced watchdog files for both a success and fail

The CR100x system as shown above is using a port forward via a Cybertec modem. This setup is working as expected on both cell and LAN connections. Watchdog code and logger setting xml in dropbox file

The CR300 Cell200 is the one that is not sending Email. the code will ping our exchange server, I can IP connect to the device.. but when the Email code executes. it fails.. and the modem reboots the cell connection..

The Email Result is hung on   "220 2.0.0 SMTP server ready " prior to the fail and subsequent reboot

similar watchdog and code included

Curious to know what is going on and where the fault is.. happy to undertake further tests or provide additional information if required

regards

Paul

Log in or register to post/reply in the forum.